Storing duration Temperature Rapid Test
ACCURATE ® test device ACON ® One Step Test Device
Sensitivity % Specificity % Sensitivity % Specificity %
2 weeks 37°C 96 100 96 100
45°C 96 100 96 100
50°C 96 100 96 100
4 weeks 37°C 96 100 96 100
45°C 96 100 94 100
50°C 96 100 92 100
8 weeks 37°C 92 100 92 100
45°C 92 100 90 100
50°C 84 100 84 100
Table 3: Sensitivity and specificity of DBS stored at 37°C, 45°C, and 50°C tested with the two rapid tests after storage for two, four and eight week.