Storing duration
Temperature
Rapid Test
ACCURATE
®
test device
ACON
®
One Step Test Device
Sensitivity
%
Specificity
%
Sensitivity
%
Specificity
%
2 weeks
37°C
96
100
96
100
45°C
96
100
96
100
50°C
96
100
96
100
4 weeks
37°C
96
100
96
100
45°C
96
100
94
100
50°C
96
100
92
100
8 weeks
37°C
92
100
92
100
45°C
92
100
90
100
50°C
84
100
84
100
Table 3
: Sensitivity and specificity of DBS stored at 37°C, 45°C, and 50°C tested with the two rapid tests after storage for two, four and eight week.