| Feedback devices | The effect of    using a CPR feedback device (with vs. without) | 
      
        | Compression depth | Compression rate | 
      
        | Depth
          (mm)* | No. at “correct”
          depth (%)* | P-value | Compressions
          (rate minute-1)* | No. at “correct”
          rate (%)* | P-value | 
      
        | Voice-advisory manikins (VAM) | 
      
        | “Skill Reporter”19 | 40.0 vs. 36.7 |  | 0.018 | 106.7 vs. 106.0 |  | ns | 
      
        | Motion capture technologies (or virtual reality enhanced manikins,    VREM) | 
      
        | Infra-red camera21 | 69.0 vs. 65.0 |  | ns | 121.5 ± 11 vs.    102.6 ± 18 |  | <0.001 | 
      
        | Microsoft Kinect™(mini VREM)22 |  | 47.4 vs. 24.9 | 0.002 |  | 72.0 vs. 31.4 | <0.001 | 
      
        | Defibrillators with built in feedback functions | 
      
        | Phillips    HeartStart-Mrx™ with Q-CPR™ 26 | 38 ± 6 vs. 34 ± 9 | 53 vs. 24 | <0.001 | 109 ± 12 vs.    121 ± 18 |  | 0.001 | 
      
        | Investigational    defibrillator
          (IDE #G020121)27 | 44 ± 10 vs. 43 ± 12 |  | 0.47 | 100 ± 13 vs.    104 ± 18 |  | 0.16 | 
      
        | Portable feedback devices | 
      
        | CPREzy™23 |  | 71.2 vs. 34.1 | ≤0.01 |  | 93.7 vs. 19.8 | ≤ 0.01 | 
      
        | PocketCPR™24 | 50.8 ± 10 vs.    38.1 ± 10 |  | 0.05 | 101 ± 9.7 vs.    127 ± 13.8 |  | <0.001 | 
      
        | CPRmeter™25 |  | 73.1 vs. 45.2 | <0.001 |  | 94.6 vs. 62.4 | < 0.001 | 
      
        | Smartphone    applications | 
      
        | iCPR™31 | 37.2 ± 12 vs.    41.1 ± 13 |  | 0.57 | 101.1 ± 3 vs.    107.8 ± 21 |  | <0.01 | 
      
        | PocketCPR™32 | 40.0 vs. 41.0 |  | 0.85 | 92.8 vs. 96.1 |  | 0.57 | 
      
        | CPR=Cardiopulmonary resuscitation, VAM =    voice advisory manikins, VREM=virtually enhanced manikins *data presented as intervention    group  vs. control group
 |