Method

NO/NO2- Detection Limit*

Continuous or Discrete?

Direct Measure in Blood?

Lab Sample Prep Required?

Potential as rapid clinical diagnostic?

EPR

~10 nM [139]

Continuous

Yes

No

No

MRI

~1 mM [54]

Continuous

Yes

No

No

Griess Reaction

10 -1500 nM [63]

Discrete

No

Yes

No

Ozone-Chemiluminescence

1 nM [63]

Discrete

No

Yes

No

Ion-selective nitrite electrode

0.5-10 µM [70]

Continuous

Yes

No

Yes

Nitric oxide electrode

0.08-1 nM [55]

Continuous

Yes

No

Yes

HPLC

10 pM [68]

Discrete

No

Yes

No

Capillary Electrophoresis

60 nM [69]

Discrete

No

Yes

No

GC-MS

0.5 µM [28]

Discrete

No

Yes

No

Exhaled NO

0.5 ppb (~ 15 nM) [140]

Continuous

No

No

No

Exhaled breath condensate

~ 1 µM [61]

Continuous

No

No

No

Silver Medical instrument

< 10 nM [71]

Continuous

Yes

No

Yes

Table 1: Comparison of current NO and NO2 - detection methods. *Detection limits are approximate since these can vary in practice.