Overview of Validation Exercises and Criteria
Validation Exercise Minimum
Experiments
Performance Criteria
Selectivity Matrix blanks: 6 lots, n=1 for each lot
Matrix blank fortified with IS: 6 lots, n=1 for each lot
LLOQ Selectivity Sample: 6 lots, fortified with analyte at LLOQ level and IS. n=3 for each lot
At least 5 out of the 6 lots must meet the following criteria: Response for the analyte in matrix blanks or matrix blank fortified with IS must be £20% of the mean analyte response in the acceptable LLOQ calibration standards
Response for IS in matrix blanks must be ≤5% of the mean IS response in the acceptable LLOQ calibration standards
At least two-thirds of the selectivity LLOQ replicates for each lot must meet accuracy acceptance limit, and the mean accuracy must be within ±20.0% of the nominal concentration
Cross-analyte Interference Each analyte at ULOQ evaluated separately. IS at the level of use evaluated separately Interference must be £ 20% of the mean analyte peak response or £ 5% of the mean IS response of the acceptable LLOQ calibration standards
Carryover Minimum of 2 blank samples after the injection of each ULOQ standard Analyte response £20% of the mean analyte response in the acceptable LLOQ standards, IS response ≤5% of the mean IS response in the acceptable LLOQ standards
Linearity Minimum of 6 non-zero calibration standard (CS) levels (R2) ≥0.985
Calibration
Standards:
Accuracy
Injected at the beginning and end of the analytical run Minimum 6 non-zero (or 75% of total) CS must be within ±15.0% RE of nominal (exception: LLOQ within ±20.0 %RE)
QC Samples –
Core Validation
Three concentration levels: Low, Mid, High; n=6 at each level Minimum 50% of the QC replicates at each level and 66.7% of all QCs must be within ±15.0% RE of nominal
Mean inter- and intra-assay accuracy within ±15.0% RE of nominal; precision £15.0% RSD
QC Samples –
Ancillary Runs
Three concentration levels: Low, Mid, High; n ≥ 2 at each level ≥66.7% of all QCs and at least 50% at each level must be within ±15.0% RE of nominal
LLOQ Samples
(Sensitivity)
n=6, ≥ 1 run Mean accuracy within ±20.0%RE of nominal; precision £20.0% RSD
Recovery Analyte at low, medium and high levels, and IS at the level of use: pre-extraction spiked samples (n=6) are compared with mean response of post-extraction spiked matrix samples (n=6) Recovery for analyte and IS must be
relatively consistent across all QC levels
Dilution (Parallelism) One level (minimum 10 fold dilution); n=6 Mean accuracy within ±15.0% RE of nominal; precision £15.0% RSD
Matrix Effect Post-extraction spiked samples (n=6, at each QC low, mid and high level) are compared with mean response of 6 injections of analyte or IS in solvent MF will be calculated and reported for the analyte and for the IS
Ruggedness Minimum of two variables over the course of validation (e.g. different column, instrument and/or analyst) Mean inter- & intra-assay accuracy within ±15.0% RE of nominal; precision ≤15.0% RSD
Table 2: Overview of Validation Exercises and Criteria.