Parameter Impurity-A Impurity-B Impurity-C Impurity-D Impurity-E Impurity-F
System suitability
RT 17.375 12.714 10.539 31.270 18.380 20.544
RRT 3.174 2.323 1.920 5.713 3.358 3.753
Linearity (r2) 0.99876 0.99966 0.99937 0.99900 0.99969 0.99814
Quantitation
Limit (%) 0.013 0.008 0.012 0.009 0.003 0.0011
Detection
Limit (%) 0.003 0.002 0.003 0.0022 0.0007 0.0002
Precision at
QL (RSD) 0.74 2.39 2.26 4.21 1.47 3.43
% Recovery at
QL (n=3) 100.12 101.45 99.02 104.55 100.73 98.09
Accuracy (% Recovery)
50% 100.61 99.11 103.14 98.62 101.05 97.86
75% 102.14 99.31 99.39 99.59 94.23 98.29
100% 97.25 100.18 99.90 98.48 101.23 103.42
125% 92.72 99.28 98.20 98.49 101.60 106.90
150% 100.66 98.44 98.83 99.68 98.44 105.80
Note: n, number of determinations; RT, retention time; RRT, relative retention time; R2, correlation coefficient. Table 1 shows different parameters of validation along with their respective results. It includes system suitability, linearity, quantitation limit, detection limit, precision at QL, and accuracy (with various concentrations such as 50%, 75%, 100%, 125%, 150%).
Table 1: Method validation summary report.