Iron (Fe) m/z |
57 |
Germanium (Ge) m/z |
72 |
Radio frequency (RF) power |
1.3 kW |
Argon (Ar) cool gas flow |
Typically 16 L/min |
Ar plasma gas flow |
Typically 0.80 L/min |
Ar nebulizer gas flow |
Typically 0.8 – 1.1 L/min |
Nebulizer |
PFA-100 micro flow |
Spray Chamber |
Cyclonic |
Detector mode |
Dual |
Measurement units |
cps |
Peristaltic pump speed |
12 R/min |
Peristaltic Pump Tubing |
Black-black (0.76 mm i.d. x 150 mm 2-stop, PVC) |
Sample uptake time |
120 s |
Wash time |
120 s |
Resolution mode |
Medium |
Sample time |
20 ms |
Samples/peak |
30 |
Mass Window |
300 (57Fe and 72Ge) |
Search Window |
60 (57Fe), 10 (72Ge) |
Integration Window |
20(57Fe), 20(72Ge) |
Acquisition points |
10 |
Scan type |
EScan |
*The parameters were adjusted to optimize system response on a run-by-run basis. |
|