Low Frequency RF Scanning (RF Surgical Systems Inc) High Frequency RFID
Scanning (ClearCount Medical Solutions)
Manuel Barcode Scanning (SurgiCount Medical, Inc)
Published peer reviewed studies 4 [31-34] 1 [36] 2 [19,37]
Sensitivity 100% [31] 100% [36] Not reported
Specificity 100% [31] 100% [36] Not reported
Bariatric subjects Yes [32] No No
Reduction in procedure time studied Yes [34] No No [19]
Cost Savings over Implantation Costs Yes [34] No No
Reported Staff Satisfaction with system 97% [33] 85-93% [36] 60% [37]
Avoidance of biohazard material prior to scan Yes [28] Yes No
Diameter of Detection wand 14’’ [28] 21’’ [40] NA
In-Vivo sponge detection ability Yes [28,33] Yes [29] No
Validated performance in tissue and fluids Yes [28] No [39] No [30]
Operative Field sponge detection ability Yes [28,33] Yes [29] No
Effective through bone cement Yes [28] Yes No [38]
Numbers correspond to listed citations
Table 3: Comparative Review of FDA Cleared Technologies for Prevention of retained Foreign Items Source: www.Rfsurg.com.
Goto home»