Sample Identity

Negative control

Liver in L-15 medium

ISAV 101 in liver homogenate

ISAV 102 in liver homogenate

ISAV 105 in liver homogenate

ISAV 103

in L-15 medium

ISAV 104

in L-15 medium

Positive control (ISAV 106 in L-15 medium)

Sample Codes

Lab’s own

01, 0B, 0D, 012, 0M, 018

0A, 0G, 0E, 015, 0T, 0K

08, 011, 0C, 0L, 0Q, 016

0H, 05, 07, 019, 017, 013

04, 06, 02, 0R, 0J, 0P

09, 0F, 03, 0N, 014, 0S

ISAV

Testing Lab

 

 

 

 

 

 

 

 

Lab A

0

42.97±0.511 (3/6)2

41.02±0.49

31.14±15.37(5/6)

29.9±0.91

35.41±0.33

32.24±0.32

25.13± 0.17

Lab B

0

0

35.98±0.34

33.67±0.68

31.8±0.53

32.46±0.98

30.62±0.78

24.14±0.41

Lab C

0

0

37.06±0.99

34.2±0.33

28.41±1.4

32.0±0.23

28.94±0.26

22.34±0.19

Lab D

0

0

39.15±0.69

35.62±0.62

30.32± 1.3

34.26±1.0

30.91±0.4

24.16±0.89

Lab E

0

0

35.84±0.56

33.97±0.39

30.81±1.8

33.95±0.56

30.74±0.4

23.82±1.1

Lab F

0

0

30.9 (1/6)

31.4±1.0

24.9±0.99

29.5±0.98

26.18±0.77

19.5±0.28

Lab G

0

0

0

32.0± 0.4 (2/6)

28.89±1.1

31.41±0.37 (2/6)

31.04±0.99

25.98±1.93

Lab H

No report

No report

No report

No report

No report

No report

No report

No report

Lab I

0

0

38.85±0.62

36.18±0.86

28.13±0.62

34.39±0.79

31.28±0.62

23.72±0.78

Lab J

0

0

37.82±1.3 (5/6)

35.22±1.61

28.04±1.18

33.3±1.45

30.14±1.4

24.63±0.98

Lab K

0

0

40.87±1.1

36.85±0.77

30.0 ± 0.59

35.11±1.03

31.78±0.64

24.09±0.44

Lab L

0

0

0

31.11±0.66

26.59 ±1.18

29.39±0.27

26.58±52

20.58± 0.58

Lab M

0

0

29.64 ±2.17 (3/6)

29.06 ±1.69

27.05 ± 3.64

28.96 ±0.83

27.49 ±0.45

21.05±1.1

Lab N (AB)

40

39.9±0.2

36.39±1.2

33.57±0.45

24.2 ± 0.59

31.23±1.4

28.08±1.43

22.1±0.11

Lab O (Denmark)

0

0

0

38.12±0.60

29.04±0.85

34.41±0.60

31.76±0.87

25.2±1.27

OIE Ref Lab, AVC

0

0

34.98±0.39

31.97±0.38

31.27±0.64 (3/6)

32.55± 0.58

29.63±0.38

22.7±0.55


Mean Ct value ± SD1 for assays performed on 6 different days. Where a sample had no Ct value (i.e., Ct = 0), the number of days in which it had a Ct value (Ct = >0) out
of the six days it was tested is given in brackets2.
In Blue: denotes mean Ct values that are ≥3 more than expected.
In green: denotes unexpected false negative(s).
In magenta: denotes sensitivity of real-time RT-PCR for ISAV is below that of most laboratories.
In red: denotes poor performing laboratory based on having more than two days (a) with mean Ct values that are ≥3 more than expected, and/or (b) with unexpected
false negative(s)
, and/or (c) unable to detect a virus titer of 103 TCID50/ml.
General Observations:
1. Cut off Ct values range between≥40 and ≥35 = negative.
2. No lab had a false positive result.
3. Lowest amount of virus titer detected by most laboratories is 103 TCID50/ml.
4. The best performing laboratories are: Lab B and Lab E.
Table 3: Overall performance (inter-lab and inter-assay) of real-time RT-PCR for ISAV with TaqMan segment 8 probe (Snow et al. 2006) in Ring Test Phase II.